I'm upgrading a 24v Permobil F3 power wheelchair to 16 100Ah LFP cells. https://www.expertpower.us/products/16-3-2v-100ah-lifepo4-cells because they are a perfect fit in the integral battery box/vehicle frame. I'd like to divide the 16 cells between 4 4S active cell balancers. Creating 4 12v cell groups that can be rapidly reconfigured as 1S4P 12v, 2S2P 24v, 4S1P 48v From here on, I won't refer to individual cells. Only 4 4S cell groups
It is the chair I only use outside of the house. It spends most of its time in the van. I want it to charge rapidly as possible from the van's 270A alternator. I already have a Foval 60A/30A 12vdc-12vdc multichemistry charger. But ideally the chair could also charge from any running 12v vehicle with only onboard equipment and limit its draw to <10A to preserve the cigarette lighter fuse. 12v stationary charging configuration 1S4P will also be used to discharge to external DC loads like CPAP, Starlink mini, laptop, 1000W inverter. And to jump start the van engine in a pinch.
Normal 24v operation w/brushed motors requires 2S2P. These particular cells have reports of swelling even at max 0.5C(50A) charge and max 1C(100A) discharge. The Roboteq motor controller can deliver up to 150A x 2 to motors. So it will be important to monitor the discharge current as read by the active balancers, and reduce motor controller PWM duty as limits are approached. I cannot have a BMS that abruptly cuts off the load. It is a life safety issue. Even if it might kill the batteries, I need it to keep delivering whatever current it can (but set a warning) in case I happen to be half way across the street or similar precarious place when the BMS sees something it doesn't like. If only a 24V mobility charger is available, as may happen in a hospital or other care facility, that could be used in this configuration.
High speed 48v operation w/overvolted 24v brushed motors 4S1P. This is a dangerous experimental mode and the chair is very likely to become unstable, fishtail, and maybe flip. Initial tests will be remote controlled and used to fine tune response to gyro. Definitely no sudden battery shutdowns allowed. Have 4 100w bifacial solar panels and Tristar 30A MPPT that can charge this configuration in a power outage or off grid camping.
What are the best active balancers for this kind of highly versatile application? I want to monitor and log individual cells. CAN, RS485, Bluetooth, I don't care as long as I have fine grained continuous monitoring which I can log to NVME with a Raspberry Pi 5 onboard.
Was considering 4 ENJBMS2A4S200SP 8 series/unlimited parallel active BMS. I'm finding very limited information about these BMS and what is necessary to rapidly reconfigure them. Their contact form fails. I've heard that almost any BMS can be used this way as long as you do not use P- or C-. Just B- and balance connectors. How true is that? On hand, I have a single JK-B2A24S20P-HC to get me through initial 48v testing.
What do you all think is the best way to reconfigure the wiring between 1s4p for 12v charging, 2s2p for normal 24v, 4s1p for 48v high speed and 48v solar charging? I've got 8 wires to switch around. I could make up a set of Andersen PP75 connectors for each of the 3 configurations. Would that be enough? Does anything else have to happen before pulling one plug and inserting another? I don't think I could separate or mate 8 PP75 connections simultaneously. Same terminal as SB50 in red/black modular housings like PP45 only larger. What other options are there? Contactors, MOSFET switches? Too much parasitic drain. Maybe some kind of rotary switch. A drum switch?
I'm sure I've made mistakes. I don't have much experience with BMS at all. I read this in a JK manual
"ESS BMS Current limiting function:
1.The current limiting function is mainly used to balance the voltage between battery packs when connected in parallel;
2.The current limit is 10A;"
I'm not quite sure what that is, but was wondering if I can use to limit the current drawn from lighter socket.
It is the chair I only use outside of the house. It spends most of its time in the van. I want it to charge rapidly as possible from the van's 270A alternator. I already have a Foval 60A/30A 12vdc-12vdc multichemistry charger. But ideally the chair could also charge from any running 12v vehicle with only onboard equipment and limit its draw to <10A to preserve the cigarette lighter fuse. 12v stationary charging configuration 1S4P will also be used to discharge to external DC loads like CPAP, Starlink mini, laptop, 1000W inverter. And to jump start the van engine in a pinch.
Normal 24v operation w/brushed motors requires 2S2P. These particular cells have reports of swelling even at max 0.5C(50A) charge and max 1C(100A) discharge. The Roboteq motor controller can deliver up to 150A x 2 to motors. So it will be important to monitor the discharge current as read by the active balancers, and reduce motor controller PWM duty as limits are approached. I cannot have a BMS that abruptly cuts off the load. It is a life safety issue. Even if it might kill the batteries, I need it to keep delivering whatever current it can (but set a warning) in case I happen to be half way across the street or similar precarious place when the BMS sees something it doesn't like. If only a 24V mobility charger is available, as may happen in a hospital or other care facility, that could be used in this configuration.
High speed 48v operation w/overvolted 24v brushed motors 4S1P. This is a dangerous experimental mode and the chair is very likely to become unstable, fishtail, and maybe flip. Initial tests will be remote controlled and used to fine tune response to gyro. Definitely no sudden battery shutdowns allowed. Have 4 100w bifacial solar panels and Tristar 30A MPPT that can charge this configuration in a power outage or off grid camping.
What are the best active balancers for this kind of highly versatile application? I want to monitor and log individual cells. CAN, RS485, Bluetooth, I don't care as long as I have fine grained continuous monitoring which I can log to NVME with a Raspberry Pi 5 onboard.
Was considering 4 ENJBMS2A4S200SP 8 series/unlimited parallel active BMS. I'm finding very limited information about these BMS and what is necessary to rapidly reconfigure them. Their contact form fails. I've heard that almost any BMS can be used this way as long as you do not use P- or C-. Just B- and balance connectors. How true is that? On hand, I have a single JK-B2A24S20P-HC to get me through initial 48v testing.
What do you all think is the best way to reconfigure the wiring between 1s4p for 12v charging, 2s2p for normal 24v, 4s1p for 48v high speed and 48v solar charging? I've got 8 wires to switch around. I could make up a set of Andersen PP75 connectors for each of the 3 configurations. Would that be enough? Does anything else have to happen before pulling one plug and inserting another? I don't think I could separate or mate 8 PP75 connections simultaneously. Same terminal as SB50 in red/black modular housings like PP45 only larger. What other options are there? Contactors, MOSFET switches? Too much parasitic drain. Maybe some kind of rotary switch. A drum switch?
I'm sure I've made mistakes. I don't have much experience with BMS at all. I read this in a JK manual
"ESS BMS Current limiting function:
1.The current limiting function is mainly used to balance the voltage between battery packs when connected in parallel;
2.The current limit is 10A;"
I'm not quite sure what that is, but was wondering if I can use to limit the current drawn from lighter socket.